What criteria can be defined for system suitability in UV Methods for Assay, Impurities (Limit test)

Methods, Validations

All Replies

Viewing 1 reply thread

  • Up
    0
    Down
    ::

      Pharmacopeia do not define specific system suitability tests for UV methods. However the following may be considered.

      1. A test of repeatability by measurement of response of a standard solution, for example response of standard in five replicates (with %RSD to be less than 2.0%)
      2. Resolution with a solution of toluene in hexane,  criteria – ratio between the minimum and maximum absorbance at respective wavelengths of 267 and 269 nm must exceed 1.3 (USP).

      Useful references:

    • Up
      0
      Down
      ::

        Help Shape the Future of Packaging Inspection in Pharma & Biotech

        We’re conducting a targeted industry survey on packaging inspection methods in the pharmaceutical and biotech sector.

        Biotec is an advanced automated inspection method that captures and analyzes high-resolution images of every unit.

        Why Participate?
        Your insights will directly inform industry best practices in packaging QA and risk management. As a thank-you, all participants will receive a complimentary report benchmarking their inspection processes against peer organizations—so you can see exactly where you stand and where you can improve.