Pharmacopeia do not define specific system suitability tests for UV methods. However the following may be considered.
A test of repeatability by measurement of response of a standard solution, for example response of standard in five replicates (with %RSD to be less than 2.0%)
Resolution with a solution of toluene in hexane, criteria – ratio between the minimum and maximum absorbance at respective wavelengths of 267 and 269 nm must exceed 1.3 (USP).
Pharmacopeia do not define specific system suitability tests for UV methods. However the following may be considered.
Useful references: